oscilloscope Help!

I have recently bought a used 465m tektronics oscilloscope.

I am looking for help on calibration and or advice on if its working or not, and if not how much to repair or is it worth it.

when I hook the probe up to the calibration signal in the front of the unit I get a standard digital wave form but it looks more sawtooth than square.

The adjuster on the probe itself allows me to adjust it a little but not enough to make it looke like a conventional square wave.

just to be sure the generator is working I have hooked it up to the parallel port of a computer when mach3 was running the spindle and the sawtooth pattern is still there if not more pronounced..

Is the scope out of calibration and if so is it a DIY fix or do I have to send it out. I only payed 100.00 U.S. for it and if its fixable at a resonable price I will get it fixed. but would like to DIY.

Any suggestions???
__________________
thanks
Michael T.
"If you don't stand for something, chances are, you'll fall for anything!"
Reply With Quote

  #2   Ban this user!
Old 09-12-2005, 07:24 AM
gar gar is offline
 
Join Date: Mar 2005
Location: USA
Posts: 1,498
gar is on a distinguished road

050912-0704 EST USA

miljnor:

On any Tek scopes I have the calibration signal is a square wave.

Bypass the probe and connect a Y input to the calibration signal and see if the waveform changes. This will eliminate the probe as a cause. Check all Y channels. My 7603 has 4.0, 0.4, 0.04 outputs.

Service from Tektronix is very expensive.

The scope might be fine except for the calibrator. I doubt that the calibrator was designed as a sawtooth generator. One of the uses for the calibrator is to do high frequency compensation of the probe. The square corner of a square wave is best for this.

To check the basic operation of the scope there are other procedures you can use.

.
Reply With Quote

  #3   Ban this user!
Old 09-12-2005, 10:13 AM
 
Join Date: Aug 2005
Location: USA
Posts: 1,622
One of Many is on a distinguished road

A couple things to try:

Make sure the outlet this scope is plugged in to is properly wired for Hot, Nuet, and ground. You should also be using an adapter to go form 3 prong to 2 prong so as to isolate the earth ground from the units cord. Hopefully what you are seeing is not a result of this very condition by the previous user.

You can use a simple jumper wire to go from either channel and connect directly to the cal terminal. The grounding should be internal, so no need for the ground clip. Another device with a square wave might help verify the input or cal signal as the faulty portion.

The probe you have may not be a good match for the input. 10-60 Pf would be reasonable, but some of the cheaper probes can be in the 180pf range. Which can contribute to a slow rise time viewing the signal and never adjust to a see a good sharp leading edge to a square wave.

I'd expect any repair facility will need to go through the unit as a whole. I don't know of any that would just fix one issue. The quote I had for inspecting mine 15 years ago wanted around $900-1200 to inspect, repair and calibrate. Not worth it on an obsolete scope used for basic hobby applications. Far cheaper to find a known good scope at the $500 rate.

DC
__________________
Learn cause and effect through experience. Mastering those relationships is the "Common Sense" ability within the art of any trade.
Reply With Quote

  #4  
Old 09-12-2005, 10:16 AM
miljnor's Avatar
S.N.A.F.U.
 
Join Date: Jan 2005
Location: usa
Posts: 1,844
miljnor is on a distinguished road

Bypass the probe and connect a Y input to the calibration signal and see if the waveform changes. This will eliminate the probe as a cause. Check all Y channels. My 7603 has 4.0, 0.4, 0.04 outputs.

where are thes outputs as it has been 22 years since I have used a scope . there are several connections in the rear of the scope and of course there are several in the front. 2 in the front are singnal 1 and 2 then on the right hand side there are 2 more all are listed as 20pf and 1m on the front. the ones in the back have various listings.

If this fixes the pattern then I would assume that the probe (which is new) isn't compatable with the scope. OR more likely there is something wrong with the scope. how would I fix it??? Is this something you know how to do?
__________________
thanks
Michael T.
"If you don't stand for something, chances are, you'll fall for anything!"
Reply With Quote

  #5   Ban this user!
Old 09-12-2005, 11:20 AM
 
Join Date: Aug 2005
Location: usa
Posts: 48
EDM_Fixer is on a distinguished road
scope

i don't work for them but...... radio shack has a $69.00 i think .. meter.. that also comes w/ 232 port ..with the cable.. w/software and its a scope also .. looks like it would be great for someone starting out ..it even has the yellow rubber bumper cover .. heck i payed $35.00 for the cover alone on my fluke $2,200.00 for my scope.. ahh the speed electronics are advancing.... have fun .. Jim
Reply With Quote

Sponsored Links
  #6  
Old 09-12-2005, 12:59 PM
miljnor's Avatar
S.N.A.F.U.
 
Join Date: Jan 2005
Location: usa
Posts: 1,844
miljnor is on a distinguished road

originaly posted by OneOfMany
You can use a simple jumper wire to go from either channel and connect directly to the cal terminal. The grounding should be internal, so no need for the ground clip. Another device with a square wave might help verify the input or cal signal as the faulty portion.
direct wired made the square wave practically purfect! So this means that the probe I have is not compatable??? It is a 150mhz probe with a x1/x10 setting and the 18pf inductance cap. Or is it that the probe just has too much resistance for the 1volt 1khz Calibration signal???

what probes would you guys recomend?

originaly posted by OneOfMany
Make sure the outlet this scope is plugged in to is properly wired for Hot, Nuet, and ground. You should also be using an adapter to go form 3 prong to 2 prong so as to isolate the earth ground from the units cord. Hopefully what you are seeing is not a result of this very condition by the previous user.
what problems would the ground wire cause. I have a cheap home depot grand tester and it says the ground is good at the plug. should I still isolate the ground prong? and if so why if it goes to a good earth ground?
__________________
thanks
Michael T.
"If you don't stand for something, chances are, you'll fall for anything!"
Reply With Quote

  #7   Ban this user!
Old 09-12-2005, 02:33 PM
 
Join Date: Aug 2004
Location: US
Posts: 2,786
ViperTX is on a distinguished road

Check if you are AC coupled?
Reply With Quote

  #8   Ban this user!
Old 09-12-2005, 02:59 PM
 
Join Date: Aug 2005
Location: USA
Posts: 1,622
One of Many is on a distinguished road

Originally Posted by miljnor
direct wired made the square wave practically purfect! So this means that the probe I have is not compatable??? It is a 150mhz probe with a x1/x10 setting and the 18pf inductance cap. Or is it that the probe just has too much resistance for the 1volt 1khz Calibration signal???

what probes would you guys recomend?



what problems would the ground wire cause. I have a cheap home depot grand tester and it says the ground is good at the plug. should I still isolate the ground prong? and if so why if it goes to a good earth ground?
If the voltage scale is set to 1v/div and the probe is set on 1X, The adjustment should cancel out any capacitance created in the cable verses the scope input. Inductance in general, should have nothing to do with this probe.

I would investigate what that probe actually is. It may have a problem or is not a standard passive probe? No or not enough trimmer compensation, from the sounds of it.

Tektronix Probe Tutorial

Not all tests are taken in reference to ground. When checking some circuits/components that are not ground, but some potential above ground. Sparks will fly if the ground clip is attached to anything not at ground potential!

DC
__________________
Learn cause and effect through experience. Mastering those relationships is the "Common Sense" ability within the art of any trade.
Reply With Quote

  #9  
Old 09-12-2005, 04:32 PM
miljnor's Avatar
S.N.A.F.U.
 
Join Date: Jan 2005
Location: usa
Posts: 1,844
miljnor is on a distinguished road

by ViperTX
Check if you are AC coupled?
I am alot rusty! could you explain?

And I will check the docs I have for the probe and see if the probe is right for my ap and If i cant tell then I will assume that it dosn't have enough compensation!

thanks guys
__________________
thanks
Michael T.
"If you don't stand for something, chances are, you'll fall for anything!"
Reply With Quote

  #10   Ban this user!
Old 09-12-2005, 05:12 PM
gar gar is offline
 
Join Date: Mar 2005
Location: USA
Posts: 1,498
gar is on a distinguished road

050912-1647 EST USA

miljnor:

As has been said your probe is the problem. Get new probes from Tektronix. These each will cost more than your scope, but are well designed. Get X10 probes.

I just took a look at the Tek web site and it appears useless. I can not find a way to get at specs on probes. There are too many very bad web sites like this. Bit this does not change the fact that Tek probes are good.

If you are going to unground your scope you have to be very careful. In the case of a HAAS machine I have a special socket on the HAAS that derives from the safety ground of the HAAS and this must go back to your service entrance safety ground. See some of my discussion on noise and ground on my web site. www.beta-a2.com and then to NOISE and GROUNDING page.

If you ground your scope to the machine, then use two input channels in differential mode with both amplifiers balanced then you can cancel out most noise.

.
Reply With Quote

Sponsored Links
  #11  
Old 09-12-2005, 06:41 PM
miljnor's Avatar
S.N.A.F.U.
 
Join Date: Jan 2005
Location: usa
Posts: 1,844
miljnor is on a distinguished road

AFter I got the wire method working I decide to put the probe back on for another try. This time it didn't do anything!

So, I took it back for a replacement and picked up another one of 60mhz just to check it (they have a great return policy and know me pretty well). Turns out the first probe I bought was DOA! and the both of the new ones were perfect right out of the box (or bag as the case may be).

Thanks for the info guys and as always its greatly appreciated!

Gota love this sight!
__________________
thanks
Michael T.
"If you don't stand for something, chances are, you'll fall for anything!"
Reply With Quote

  #12   Ban this user!
Old 09-12-2005, 06:43 PM
 
Join Date: Jun 2003
Location: Boalsburg PA
Posts: 844
unterhaus is on a distinguished road

Probes and test leads go bad a lot. If I'm getting bad results, that's the first thing I check. I suspect that a probe from a company like pomona will do what you want.
Reply With Quote

Reply




Currently Active Users Viewing This Thread: 1 (0 members and 1 guests)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Search
Display Modes

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off
Trackbacks are On
Pingbacks are On
Refbacks are On





All times are GMT -5. The time now is 10:37 PM.





Powered by vBulletin® Version 3.8.7
Copyright ©2000 - 2012, vBulletin Solutions, Inc.
Content Relevant URLs by vBSEO
Template-Modifications by TMS

1 2 3 4 5 6 7 8 9 10 11 12 13 14 15 16 17 18 19 20 21 22 23 24 25 26 27 28 29 30 31 32 33 34 35 36 37 38 39 40 41 42 43 44 45 46 47 48 49 50 51 52 53 54 55 56 57 58 59 60 61 62 63 64 65 66 67 68 69 70 71 72 73 74 75 76 77 78 79 80 81 82 83 84 85 86 87 88 89 90 91 92 93 94 95 96 97 98 99 100 101 102 103 104 105 106 107 108 109 110 111 112 113 114 115 116 117 118 119 120 121 122 123 124 125 126 127 128 129 130 131 132 133 134 135 136 137 138 139 140 141 142 143 144 145 146 147 148 149 150 151 152 153 154 155 156 157 158 159 160 161 162 163 164 165 166 167 168 169 170 171 172 173 174 175 176 177 178 179 180 181 182 183 184 185 186 187 188 189 190 191 192 193 194 195 196 197 198 199 200 201 202 203 204 205 206 207 208 209 210 211 212 213 214 215 216 217 218 219 220 221 222 223 224 225 226 227 228 229 230 231 232 233 234 235 236 237 238 239 240 241 242 243 244 245 246 247 248 249 250 251 252 253 254 255 256 257 258 259 260 261 262 263 264 265 266 267 268 269 270 271 272 273 274 275 276 277 278 279 280 281 282 283 284 285 286 287 288 289 290 291 292 293 294 295 296 297 298 299 300 301 302 303 304 305 306 307 308 309 310 311 312 313 314 315 316 317 318 319 320 321 322 323 324 325 326 327 328 329 330 331 332 333 334 335 336 337 338 339 340 341 342 343 344 345 346 347 348 349 350 351 352 353 354 355 356 357 358 359 360 361